increase resolution of resistance measurement
Hi, I recently received my pokit pro and was wondering if it was possible to increase the resolution of the resistance measurement. I fix medical equipment and we have to check the ground resistance on almost every PM. I tried to use my Pokit pro but the meter read 0 ohm when my fluke measured 0.2 ohm. I assume this is because the pokit pro is only displaying full increments.
Is it possible to increase the resolution to show fractions of an ohm?
Thanks,
Steve
-
Official comment
Hello Hegel & Shane,
The Pokit Pro should be able to show measurements in 0.0 Ohm format, can you send a screenshot if it isn't because that would be a bug.
What Phone OS do you guys use?
With regards to the >10mA range the issue that we have here is our reliance on the +-3 readings spec for the accuracy at the bottom of the range, If we add one more decimal point we would need to make it at least +-6 readings for that range which would be better overall but we would then need individual ratings for each range to make this clear to everyone(which is a mess), I will discuss this with the team to see what we can do.
-
I second that request. At least 1 or 2 decimals in the resistance would be very useful. I also run into that in the mA measurement. Once it hits 10mA it switches to whole numbers so there is nothing between 10, 11, 12, etc mA. In my work I could really use at least down to one decimal in the mA scale.
Perhaps just a way to manually change the range, as it is now it is fixed on auto ranging.
0 -
It seems the resistance measurement is now reading 0.0. Must have been corrected in that last firmware update. Thank you, and I appreciate that you are responding to these inquiries from the field! Appreciate your support. I'm using Android version 10 on a Samsung Note 9.
0
Please sign in to leave a comment.